Promotional price valid on web orders only. Your contract pricing may differ. Interested in signing up for a dedicated account number?
Learn More

Applied Biosystems™ S2B, Ethyl Acetate, 450 mL

Catalog No. 400854
Encompass_Preferred
Change view
Click to view available options
Quantity:
450 mL
1 product options available for selection
Product selection table with 1 available options. Use arrow keys to navigate and Enter or Space to select.
Catalog No. Quantity
40-085-4 450 mL
Use arrow keys to navigate between rows. Press Enter or Space to select a product option. 1 options available.
1 options
Catalog No. 40-085-4 Supplier Applied Biosystems™ Supplier No. 400854
Only null left
Add to Cart
Add to Cart

This reagent is intended for N-terminal protein sequencing applications.

This reagent is intended for N-terminal protein sequencing applications.

Use with other protein sequencing reagents: 20% Acetonitrile in Water, Procise™

Sequencing System Model 491 and Model 494, 5% PITC in N-Heptane, Trifluoroacetic Acid, N-Butyl Chloride, 3.5% THF/Water, Isopropanol and Acetonitrile

TRUSTED_SUSTAINABILITY

Specifications

Product Type Ethyl Acetate
Shipping Condition Room Temperature
Format Bottle
Quantity 450 mL
For Use With (Application) N-terminal protein sequencing applications
What can cause low lysine residues on my Procise System?

The most common cause is aged S2B (ethyl acetate). Trace contaminants can develop over time and these seem to preferentially reduce lysine yields; more significant contamination will also reduce methionine and hydrophobic amino acid yields, and produce an overall reduction, sometimes quite significantly, in repetitive yield.

If the lot number indicates that the S2B bottle is close to a year old or more, it should not be installed. You can either replace the S2B or add high-purity DTT (dithiothreitol) to the S2B to correct this problem. The recipe for this calls for weighing out 4 mg of DTT, dissolving it in 1 ml of ethyl acetate (S2B), and then adding approximately 225 ul of this solution per full, 450-ml bottle of S2B.

In other cases, old R4 with oxidized DTT can also cause low lysine yields, and in this case such reduced yields can also be seen in the standards.

Why am I unable to sequence Tryptophan?

Tryptophan is one of the most difficult amino acids to sequence since it is readily oxidized. Oxidized Tryp (W) does not sequence. Often times you will see a derivative of W, kynuranine, which elutes between DPTU and W. It can account for 0-100% of the W signal. Replace the S2 bottle to improve recovery of Tryptophan. An alternative is to add DTT (an anti-oxidant) to S2 to increase the yield of W, or other forms of anti-oxidant can be added to R3. If you do add DTT to S2 (ethyl acetate) you may see a PITC-DTT adduct peak eluting close to Proline (P).